Wiring HMP155


ariffu Nov 23, 2010 09:35 AM

Dear,

I have been connected HMP155 sensor and logger CR1000, but the value is NAN
is the wiring cable of campbell and vaisala different ??

wiring cable from Campbell are :
1. yellow
2. -
3. white
4. blue
5. -
6. -
7. red
8. black

wiring cable from vaisala are :
1. white
2. brown
3. green
4. yellow
5. grey
6. pink
7. blue
8. red


IslandMan Nov 25, 2010 11:31 AM

Did you purchase the sensor from Campbell or Vaisala?


ariffu Nov 26, 2010 09:57 AM

Dear Dave Adams

I have been purchase the HMP155 from Vaisala and the spesification are :

- Type HMP155 Passive output version
- Temp output:
1. PT100
3. PT100
4. PT100
5. PT100

- RH output 0 - 5 Volt

Colud you help how the wiring and the programming
Thank You

ariffudin


IslandMan Nov 26, 2010 11:20 AM

The Campbell version has voltage outputs for both temperature and humidity, hence the simpler wiring. You have selected a model with the Humidity as voltage and the temperature measurement using the PRT100 sensor in it's raw form which requires more user input.
Simple solution:
Return to Vaisala and get the voltage output version.

More complex solution:
You will need a 3WPB100 or 4WPB100 to connect the PRT100 sensor to the logger. http://www.campbellsci.com/4wpb100. Page 21 and 22 of the Vaisala HMP155 manual show the wiring for the sensor and the PT100 sensor wiring configuration. Using the Vaisala and 4WPB100 manuals the wiring is simple to figure out.
Look in CRBasic for the PRT programming instruction and use the example from there.

You could also contact one of the many consultants and integrators to provide the programming wiring as well.
http://www.campbellsci.com/consultants

* Last updated by: IslandMan on 11/26/2010 @ 4:20 AM *