Clock problems


mwassmer Jun 1, 2010 03:12 PM

The clock in my CR1000 datalogger is about 2 minutes fast each day even though the lithium battery is a healthy 3.45 V. The CR1000 specs indicate that the clock should be accurate to +/-3 min/year, so something is very wrong. Do you know what the cause of this may be?

Thanks,
Mike


Dana Jun 1, 2010 07:10 PM

In cases such as this, there's usually something or someone behind the scenes setting the clock.

Do you work in an organization where all PCs are automatically synchronized to a central clock (and perhaps that clock is off 2 minutes?)?

You can check the trans*.log in the c:\campbellsci\loggernet directory. If a clock set is occurring, it will show up in the logs (e.g., you will find the words "Clock Set" with a date & time it occurred).

Regards,

Dana W.