Hi,
I have 3 wind monitors (05103-45), connected to the datalogger throught SDM-INT8, and one CSAT3. The SDM-INT8 and CSAT3 connect to a CR5000.
I'm having 2 problems:
1) I'm getting interference in the wind speed measurements when the wind monitors are stopped. i.e.: The value of wind speed should be 0, but it shows crazy values.
Without the CSAT3, the interference appears when I measure the wind direction using 60Hz integration. I get no interference using a 250us integration or when I don't use the BrHalf instruction.
Using the CSAT3, I get interference no matter what value I use for the integration. Again, I get no interference whithout BrHalf.
This is what I'm using for the wind direction:
BrHalf (WinDir_A,1,mv5000,23,Vx2,1,5000,True ,0,250,355,0)
BrHalf (WinDir_B,1,mv5000,24,Vx2,1,5000,True ,0,250,355,0)
BrHalf (WinDir_C,1,mv5000,25,Vx2,1,5000,True ,0,250,355,0)
For the wind speed:
SDMINT8 (WS_ms(),2,0000,0222,0000,0222,0,1,98,0)
How do I get rid of the interference?
2) The CSAT3 scans at 10Hz, but the wind monitors scans at a 10sec interval. What is the best way to measure all the instruments?
Thanks,
Caio